Does the POPD merit function operand  using data = 1 (system efficiency for fiber coupling) take into account Fresnel losses, performance of applied coatings, vignetting , and internal material transmission? Also wondering if these are accounted for  when using POPD data =  40, 41, 42 (fraction of total power within a defined radius).
If the apodization is set to gaussian do the vignetting diagram and transmission analysis (in polarization drop down menu) take the gaussian nature of the ray distribution into account?
Trying to find best method for determining transmission (by surface) for a gaussian beam through the system taking into account Fresnel losses, applied coatings, vignetting, and internal material transmission.
Any help is much appreciated.
island optics
Best answer by Jeff.Wilde
For calculating beam transmission per surface, I would suggest a polarization ray-tracing approach (as opposed to polarized POP) based on the Geometric Image Analysis tool.  This assumes that diffraction can be neglected.  (Note: Unfortunately, the Polarization -> Transmission tool only provides chief ray transmission per surface.)
For a Gaussian beam, it’s important to explicitly set the aperture apodization to be Gaussian and to also use an Apodization Factor of at least 1.5 (or larger), so that the wings of the Gaussian are not clipped.  The Aperture size can then be adjusted to properly set the diameter of the beam (e.g., see: Apodization Factor).  Here is an example using the Double Gauss sample model:
Then select the appropriate Geometric Image Analysis parameter settings:
For a single field point, just set the field size to zero.  The source should be “Uniform.” The image size should be large enough to capture the entire beam footprint at *any* location along the lens train.  “Use Polarization” should obviously be selected.  Then the beam transmission after propagation to any selected surface is reported as “Percent Efficiency” in the header.  This includes Fresnel/coating loss, material absorption and vignetting (if present).
Lastly, to streamline the process, the IMAE operand can be used to find the transmission values.  In this case, first “Save” the Geometric Image Analysis settings.  Then the IMAE will use the saved values unless overridden by any of the various input parameters to this operand (such as the surface number), as shown here:
  • Apr 21, 2022 7:05 PM UTC
Kevin Scales
  • May 2, 2022 4:58 PM UTC
Jeff Wilde’s answers are very good and I would consider them for your needs. I do want to address the POP options, as it does consider the effects you are looking at. As long as you click on the  Use Polarization check box in the General tab for POP when you save the settings for use in POPD, you will account for material and coating behaviors in the POP beam. The effect of obscurations, vignetting, surface apertures, etc will all be considered during propagation. For fractional beam power, these settings work on whatever field is at the surface in question, so whatever beam you have is what the operand measures. No assumptions about beam shape come into play at this point. The field is whatever it propagated as. You should, of course, always examine the beam carefully to make sure the results make sense, but if you have these various issues in the optical system, POP will notice them.
As an aside, I do want to point out that the apodization in the System Explorer for the aperture does not play a role in POP settings and the System aperture does not affect POP results, though individual surface apertures do.