Problem with using black box design files for Infinite conjugate microscope

  • November 18, 2022
  • 3 replies
  • 432 views

Forum|alt.badge.img
  • Fully Spectral
  • 52 replies

I am modeling an infinite conjugate microscope system using off-the-shelf designs from Thorlabs.

  1. Microscope objective: TL2X-SAP - 2X Super Apochromatic Microscope Objective, 0.1 NA
  2. Tube Lens 200mm: TTL200-A - Tube Lens, f = 200 mm, ARC: 350 - 700 nm

Individually the objective lens and tube lens designs works as expected. However, when they are put together, the image quality is far from nominal (1000 waves OPD).

I am unsure if it has got to do something with using multiple black box design files in a single zemax file or if there’s something fundamentally wrong with the combined design itself.

OPD
Image Sim
System

I have attached the zemax file with the combined microscope objective + tube lens and also the individual zemax BB files.

Any insight into what is causing the image quality to degrade by ≥ 1000 waves from nominal OPD is appreciated.

Thanks. 

Best answer by Mike.Jones

Look at your Aperture/Afocal checkbox.  Is it checked or unchecked?  Try both ways.

3 replies

Mike.Jones
En-Lightened
Forum|alt.badge.img+3
  • En-Lightened
  • 117 replies
  • Answer
  • November 18, 2022

Look at your Aperture/Afocal checkbox.  Is it checked or unchecked?  Try both ways.


Mark.Nicholson
Luminary
Forum|alt.badge.img+3
  • Luminary
  • 911 replies
  • November 19, 2022

Also, it looks like defocus. Try refocusing the inter-black box and the final back gocal length.


Forum|alt.badge.img
  • Author
  • Fully Spectral
  • 52 replies
  • November 20, 2022

Look at your Aperture/Afocal checkbox.  Is it checked or unchecked?  Try both ways.

Thank you, you were correct. This was checked in the objective black box design file.