Skip to main content
Solved

Limits of Importance sampling ?

  • May 4, 2023
  • 3 replies
  • 197 views

When my detector gets too small (size <1mm), no rays can get to it, even when I set importance sampling on the main object (scattering parameters x100 rays, source produces 1e6 rays).

Am I missing something or is there a limit where importance sampling stops working ?

Best answer by Kevin Scales

Hi Matthieu,

I can think of two possibilities right away that might be causing you an issue. First, are you shrinking the target size along with your detector. If you have a large target sphere for a small target object, you might be aiming rays roughly towards the target but still missing the actual detector in the middle.

Second, and probably more likely, when you use importance sampling the system has to scale the intensity downward for each ray to make up for the fact that a larger than normal number of rays are going towards this small selected region. As it happens, this can mean that you are triggering the minimum ray intensity threshold as set in the System Explorer. Try using a lower value and see if the rays manage to find the detector again.

If neither of these are the case, do you have a representative sample system that you’d be willing to share with the community?

3 replies

Kevin Scales
En-Lightened
Forum|alt.badge.img+1
  • En-Lightened
  • 185 replies
  • Answer
  • May 4, 2023

Hi Matthieu,

I can think of two possibilities right away that might be causing you an issue. First, are you shrinking the target size along with your detector. If you have a large target sphere for a small target object, you might be aiming rays roughly towards the target but still missing the actual detector in the middle.

Second, and probably more likely, when you use importance sampling the system has to scale the intensity downward for each ray to make up for the fact that a larger than normal number of rays are going towards this small selected region. As it happens, this can mean that you are triggering the minimum ray intensity threshold as set in the System Explorer. Try using a lower value and see if the rays manage to find the detector again.

If neither of these are the case, do you have a representative sample system that you’d be willing to share with the community?


  • Author
  • Monochrome
  • 2 replies
  • May 5, 2023

yes, I was reducing the size of the sphere as well.

It was indeed an issue with the relative intensity.

Would setting this relative intensity to 1/(number of source rays * number of scattered rays) be a general good rule of thumbs ?

Matt


Kevin Scales
En-Lightened
Forum|alt.badge.img+1
  • En-Lightened
  • 185 replies
  • May 5, 2023

I don’t know that there is a specific rule of thumb that has been worked out, but you would want to consider the number of scattered rays and the cross sectional area of your target. The number of source rays isn’t really an issue here because that setting is relative to the starting power of each ray. So if you initially have 1 Watt and a million rays, that relative measures is starting with the one millionth of a Watt. 

The other major factor besides number of scatters is the cross sectional area of your target. As it shrinks, the scattered rays used in importance sampling are reduced further in power, and this will go roughly as the square of the radius reduction. So if you lose your rays reducing the target radius by an order of magnitude, you’ll probably need to change the threshold by two orders of magnitude.